HP339A distortion analyser
Sync this is with a gain of 1000 on the A4U1 op amp. The offsets from the fundamental are 180Hz. This is the 3rd harmonic of 60Hz. The offsets are hum modulation from the power supply.
Yours is looking better but still has a ways to go.
David,
Yep, I know slowly....Looking for a time where I can spend about
4 hours on it. Otherwise I just set everything up and then I have to
take care of my toddler. Almost in her terrible two's but, she's acting
older than her age.
Anyway, what is the best way to lower those third harmonics (60Hz source
and 180Hz offsets? Are these a result of AGC and/or Demodulators on the
A4 board? Is it filterable? IN the HP tech notes they recommend putting
200 ohm resistors in front of L1 and L2 which feed each demodulator.
That would create an RC filter network but not sure if that would be in the
180 Hz harmonic region.
Switches are causing trouble for this 339A. The mode selector switch is causing signal leakage. I have to work the switch to get it cleared up. The leakage is strong enough to cause the final gain stage before the RMS detector to clip. When this happens the distortion on the monitor port goes way up. The 80 kHz filter is unusable. It completely undoes the 400 Hz filter's effectiveness. It will take some digging to figure out what the problem is there. But unless the switches are working properly then diagnostics are impossible.
In the mean time, did you see Richard's posts to me about removing them from the circuit...?
Post # 406, 440
Hope every one enjoyed their Holiday, Winterbreak, Christmas, Hanukah, etc.
SyncTronX;4165956 said:
Hope y'all had a good New Year.
One of my goals is to finish up the 339A.
The following apply to the A4 board, the removable board at the
back of the analyzer.
That was incorrect.
If you have an early Model then you removed
L3 and L4.
Then, insert 200 resistors in front of cap C22 and C31.
These then become R67---C31 and R68---C22.
In my attempt to undo all the mods I also removed C25,
the 650 uF cap shown in the post # 510 the lower schematic,
provide feedback for for U5B. This is only supposed to have
.6uA Leakage after being energized with 9 volts after two minutes.
The replacement cap that I used measured on the 100uV/A scale,
indicates 0.091MA Vdc. With probes directly attached to the cap
with the 9 VDC applied.
Searching Measuring Cap Leakage Current brought up
information using a resistor and measuring the voltage.
Not sure what the correct method is to measure it now.
This should be a standard but tricky measurement.
Increasing the challenged it is presumed the original low leakage
sprague cap was disposed of. whoop ti do.
What is installed now is a Panasonic 105C 680uf with a small
.1uf 63V whima bypassing it. Measures about 638uf.
Me thinks I"m losing all my hair over this.
Other interesting notes are to replace the .062A fuses with .1V fuses.
For less 3H at low frequencies.
These are some of the changes published by HP in Supplement A for 00339-90001, 1983. They should be found at the end of the HPs
online manual.
Pics at 11.
cheers,
Thanks Demian,
I have to look to find one. It looked like a electrolytic from Sprague,
from what I recall. Tant's of that value run 230.00 to $550.00 at mouser.
In the mean time I'm stuck with what I put in there. Almost wondering
if I could put two electrolytics back to back with a film bypass.
Not sure how that affects leakage.
Thinking what I have now is a Panasonic FC with a small wima film cap
should prevent leakage.
However,
I could use a small value cap .1uf on the
pin or in series with the feed to the
quad opamp. This should block
the DC from the causing distortion
on the opamp.
OR
Use that small caps then wire the other
side to ground ground. w/ wo a diode
to block other signals from entering
the opamp.
There is another cap that has even
lower specs. This is A2C23 for
the meter. It originally was a 10uf
cap, but now is a 15uf cap with a
maximum leakage spec of 0.05uA
after 2 minutes with 14VDC applied to it.
"If the dc leakage through C23 exceeds .05uA
a dc voltage offset occurs at pin 4 of U5 causing
and erroneous meter reading.
I'm trying, this is tricky stuff and standard parts just
don't seem to work very well. Heck I even cleaned off most of the bench
looking for that 650uf cap. A lot of tants I've seen don't
even meet that spec. The Panasonic FC I used has a spec
of 3uA leakage. : (
Now I'm wondering if I should have used switch mode supply
capacitors for the mains filters in the power supply.
...and the beat goes on.
I've found some schematics for measuring leakage current
and in general it seems that it is a black art.
Do not modify circuits whose function you don't really understand. David has been repeating this point for a while and he is correct. HP knew a great deal about making this thing work, so just follow their lead and do your best to do what they recommend and get everything working to spec before you start messing around. We all want to see you succeed, but there is really a lot you don't grok yet, and it's causing you all kinds of problems. Go back to the beginning, start over, and follow the rules, at least for now.
After being away from the 339A for awhile... I came back to it and it appears that the analyzer's minimum reading may not be noise limited but is seeing the notch frequency residual. How can the notch be made deeper... 10+dB would be good.
THx-RNMarsh
THx-RNMarsh
The notch depth can be increase by increasing the gain of A4U1.
Keep in mind that removing more fundamental may result in the harmonics having greater magnitude than the fundamental. The phase detector may lock onto a harmonic rather than the fundamental. If this happens then the opposite of what you're trying to achieve will occur. Fundamental suppression will be lost.
I needed a break also.
So I got bogged down in trying to make an octopus rig/curve
tracer. Still trying to get it to work properly. That said,
Richard, do you think that might be what is going on in
my case?
It's a great new year, wish everyone the best.
cheers,
As I recall, you changed its gain resistor to 200 Ohms (5X increase). But then there was something about scaling being off... Could the gain for proper scale be similarly reduced at the output past the filters (U4) ?
THx-RNMarsh
@y'all, no not yet.
I've made no additional changes nor mods, nor undoing since
my last post a couple of weeks ago. Been in a holding pattern.
I think I'll put a pot in there with a smaller resistor in series so that
I can change the gain of the A4U1 for now. Nothing fancy while
I fiddle with the circuit.
Some of the issues I think I'm having are feeding the
oscillator back on the A1 board.
@ Richard, what electrolytic cap are you using there
at the oscillator and JFET on the A1 board? I think it is A1C30?
Brand, Series, Value? Is that a Nichicon polarized electrolytic?
Goal: pinpoint/fix the oscillator problem.
Then I can get off the waterboard. : )
I don't remember the scaling being off from that change. The tracking was lost for the said reason. You could try a lesser gain increase. X2, 3X.
Tried just a X2 gain change and no affect on thd+n result ----- so I guess it only leaves noise. So, still need to find way(s) to lower noise.... might be circuit R's limited and not old opamp's noise. but need to test that also.
THx-RNMarsh
All the caps that have no DC on them are replaced with low Z/ESR Bipolar types. That one is 100Mfd/100Vdc 105C. But I dont see the brand name on it and dont remember, either.
THx-RNMarsh