The best gift I have ever received Tektronix 2465
I began to calibrate the 2465
I checked all the voltage on j119
I let the scoop warm up for about 20 minutes
The 10V voltage was a small deviation of 10.2V
(J119-4)
i calibrated the 10 volts in R1292
After I adjusted
Exactly to 10 volt.
All the over Voltage Were exactly according to the requirements
+10V- J119-4
+87V - J119-8
+42.4V J119-9
+15V J119-6
+5V J119-2
+5V J119-12
-5V J119-5
-8V J119-11
-15V J119-1
The scoop works perfect
The screen image is very sharp.
The frequency measurement I have done with it is very accurate.
Now my problem
The only problem I'm trying to figure out is how to calibrate
the channels are not calibrated in terms of voltage measurement.
I understand what has to be done.
Correct the offset of each channel with an external frequency generator.
The question
where is the potentiometer for the Vertical gain amplifier
for each channel offset calibration ?
( the service manual for the 2465a is Very big Almost 500 pages Still trying to figure out how to do it there
So help is welcome.
Best regards David Az
Isn't an autocal routine in this scope, something like you feed in an accurate dc voltage, and activate the cal routine, check in the menu. I had the tek advanced training on this particular model decades ago, a full two weeks of training.
Check the service manual calibration section. Their manuals were among the best with HP. The complete calibration is there...
BTW be very, very carefull with the big hybrid modules, they are certainly impossible to get now, and this scope won't work with a damaged one...
SB
So you say, you try carry out 2. Adjust DAC Ref (2010) in service manual, and J119 pin13 remains meanwhile at a fixed 15V?
yes pin 13 on j119 is 15v fix .
my tek 2465 is the 300Mhz not the b version
a5 borad is different
What I've said above, it is from 2465a manual, yours is a 2465, where it is different a little bit, have you the correct manual?
wow gyuri big big thanks you are the best
I'll try it now 🙂