Anyone repair Tektronix 24xx Scopes?

Status
Not open for further replies.
Hey gang. No time to look into this today, but I'll get to it. Off hand I can say one thing I did was crank up the sweep rate and dial off the attenuation to see whether this wide-band trace thickening in channels one and two was really noise or something more periodic. The result was fairly inconclusive, but it did look more periodic than noise. It would be funny to finally find that the vertical amp module has been oscillating all along, or maybe no so funny. I'm not sure I can nail it, but will work on it in the next couple weeks. It really is a critical factor in getting this machine to do all it can do up to 300 MHz, which is twice as fast as any of my probes! I must say that I have worked around the problem so far in that the only circuits that I've been measuring that run that fast have been digital, and were mostly compatible with the limited attenuation options of inputs 3 and 4, and that's why I've not really had to corner this problem sooner...
 
Hi All,
I would be cleaning contacts with Alcohol, nothing that will remain to gum up the works. I think that Tek is fairly specific about the use of cleaners ... Don't!

If you get any fluid into a trimmer cap, that's it. You will need to replace that capacitor. It's not worth the trouble to attempt to use a contact cleaner. If you are extremely careful (which means no spraying or pumping of chemicals into the works), you can use a no-residue cleaner that completely evaporates. This means that the oxides must be removed with paper, heavy paper (better) or similar. If you float the contaminate out, where does it go? Make certain you know exactly where anything you wash out is going to end up.

-Chris
 
I bought the deoxit contact cleaner, which evaporates completely in my experience. I'll only be using this to clean the volts/div and sweep switches, removing the faceplate board from the unit completely. I'm not sure what other method to use to clean these. A tektronix tech in Israel claims that they have seen similar symptoms in aging scopes caused by these switches, so I thought I'd see if cleaning them gets results before wasting money on a replacement faceplate board.
 
Hi Justinlwtx,
The main concern is with any fluid getting into trimmer capacitors, or wicking into any other component. You need to control your delivery of the cleaner.

Business card stock is excellent for cleaning contacts. Put the cleaner onto the stock (no ink or colours in that area) and use that to deliver the cleaner and also to gently wipe the contact surfaces. I find this very effective. The contacts in Tek 2235 'scopes are much easier this way. You can also use a glass syringe (rubber swells a lot), or a dripper. The nozzle would be equivalent to that of a syringe.

For audio switches, you can use a contact cleaning solution, then follow that up with the no residue type of cleaner. Business card stock cut to the size you need does a very neat job. Of course, you always have to use your head with delicate contacts....

Always use the least amount of force to do the job and never force contacts. Also know when to accept that some switches are too far gone to repair. It really helps to have your beer or spirits after the job is done, never before! 🙂

Wishing you success, Chris
 
I bought the deoxit contact cleaner, which evaporates completely in my experience. I'll only be using this to clean the volts/div and sweep switches, removing the faceplate board from the unit completely. I'm not sure what other method to use to clean these. A tektronix tech in Israel claims that they have seen similar symptoms in aging scopes caused by these switches, so I thought I'd see if cleaning them gets results before wasting money on a replacement faceplate board.

In the future, use isopropyl alcohol.
 