req: simple method to measure L under DC bias

Status
Not open for further replies.
I am looking for a way to determine inductance in the presence of DC current.

With anything having at least two separate windings this is relatively simple: force appropriate current with a high impedance CCS down the secondary winding and measure primary inductance with L-meter - done.

But what about a choke with a single winding ... ?

Is there a method - not too complicated and without expensive equipment ?
I am not after high precision, +/-10% accuracy would be good enough ...
 
If you already have a GP LCR-meter, the simplest method is to build a small test jig including two variable lab supplies and a few components: pic 1

This is suitable for low frequency applications, and depends on the exact variety of your LCR: some require a DC path in the DUT, and the blocking cap should be omitted, some don't care, and some require a DC path with zero offset, in which case this scheme is not usable.

However, the best method to characterize inductors is to exercise them over their operating domain: that way, you can see graphically their behavior over a range of currents.
I have built such a jig, and it incorporates additional refinements, like current limit, loss measurement, etc, but basically you need to apply a 10V step (with a function generator+MOS buffer typically) and measure the resulting current over time.
In this example, the slope is 5.6A/50µs up to ~6A, then the inductor begins to saturate, and the slope changes, thus the inductance is 10*500/6=90µH up to 6A, and it is ~halved afterwards.
That is the best method, because it takes into account the large signal inductance rather than the small signal one
 

Attachments

  • Lmeasure.jpg
    Lmeasure.jpg
    139.8 KB · Views: 279
  • Lmeasure1.jpg
    Lmeasure1.jpg
    408.4 KB · Views: 259
  • Lmeasure2.jpg
    Lmeasure2.jpg
    432.1 KB · Views: 257
@elvee:
I tried the cap to hook up my L-meter before, and it doesn't work, the cap seems to fool the meter.
The step voltage looks good, let's see whether I understand it:
- current thru inductor rises linearly when constant voltage is applied to it
I = (V*dt)/L --> L = (V*dt)/I = (10V * 50us) / 5.6A = 89 uH in your example
thank you ...

@euro21:
- basically an AC voltage devider comprised of the impedance of L1 which is Z1=2*pi*f*L1 and R1
many thanks, too.

@df96:
without knowing the poster's name or the exact key word(s) it is very difficult to search the forum and actually find what you are looking for; e.g. a search for "measure inductance" brings up nothing newer than euro21's post from 2011 ...
 
If I could remember the details I could do the search for you, but I can't so I can't. I can't even remember whether the poster used a valve or a FET to put the DC and AC into the inductor. I do remember that he measured the inductance by adding a resistive load then measuring the LF rolloff. Similarly, he could estimate the stray capacitance by measuring the HF rolloff. Try using synonyms.
 