Hi all !
How to check an output transformer, I have ohmmeter Fluke of course, and also scope, frequency generator, ect ...
1.5k/84-8-16 ohms gapped for 250ma
For PSE 300B.
I want to be sure all wiring is on the right way, no wiring in opposite phase by example, ect..
Thanks in advance.
Gilles.🙂
How to check an output transformer, I have ohmmeter Fluke of course, and also scope, frequency generator, ect ...
1.5k/84-8-16 ohms gapped for 250ma
For PSE 300B.
I want to be sure all wiring is on the right way, no wiring in opposite phase by example, ect..
Thanks in advance.
Gilles.🙂
Last edited:
The easiest way is to inject a low voltage at the secondary of the DUT (Device Under Test) from a low voltage alternate source, like a heater secondary (5 or 6.3VAC), then measure secondary (primary) voltage and phase, using both channels of the oscillo simultaneously. If the DUT becomes too hot, and low voltage appears at the primary, then it is shortcircuited or not suitable for those frequency (50/60Hz). Also, the voltage ratio will give you the turns ratio, and, squared, will give you the impedance ratio.
Last edited:
Hi,
I have published an article on the subject, see: How to test an output transformer. – tubes
although checking the phase is not mentioned in the article, but it can be done as described by Osvaldo.
Any comments to the article are appreciated.
I have published an article on the subject, see: How to test an output transformer. – tubes
although checking the phase is not mentioned in the article, but it can be done as described by Osvaldo.
Any comments to the article are appreciated.
Last edited:
Hi
on this thread
OPT Characterization
you can use a reverse check test for trafo.
You have a minimum of test equipment to make a good job.
You can omit the HV generator and connect the secondary ( in origin primary) with a resistor with the nominal value declared on specs of the trafo.
Walter
on this thread
OPT Characterization
you can use a reverse check test for trafo.
You have a minimum of test equipment to make a good job.
You can omit the HV generator and connect the secondary ( in origin primary) with a resistor with the nominal value declared on specs of the trafo.
Walter
The easiest way is to inject a low voltage at the secondary of the DUT (Device Under Test) from a low voltage alternate source, like a heater secondary (5 or 6.3VAC), then measure secondary (primary) voltage and phase, using both channels of the oscillo simultaneously. If the DUT becomes too hot, and low voltage appears at the primary, then it is shortcircuited or not suitable for those frequency (50/60Hz). Also, the voltage ratio will give you the turns ratio, and, squared, will give you the impedance ratio.
It is much safer to inject the signal into a primary winding - 5V into the OT secondary can easily produce lethal voltages on the primary and necessitate using 100x oscilloscope probes to protect the 'scope.
Inject 5V into a primary winding and measure with an DMM, no high voltages. Dont connect/disconnect with the power source powered up.
And if you are not sure which winding is which, measure DC resistance and pick the largest resistance winding as the test primary winding.
Last edited: