Old 9th February 2010, 05:18 PM   #1
fredlf is offline fredlf  United States
diyAudio Member
 
Join Date: Dec 2009
Location: Portland, OR
Send a message via AIM to fredlf
Default Heathkit IT-18 transistor tester troubles

Quote:
Originally Posted by anatech View Post
Okay, that's great! It never pays to have any doubts on your test instrument readings. It is not foolish to test the battery, that's step #1 in calibration. We actually threw out the batteries and installed new alkaline batteries. Can you measure the battery voltage inside the case where the battery clips solder on? That will ensure that there is full continuity right into the circuit. Something doesn't look right with your beta measurements.

The scales on the IT-18 are X1 and X10. Read the scale directly on the X1 scale and multiply the meter indication by a factor of 10 on the X10 scale. I normally start on the X10 scale because most transistors have beta higher than 20. So, set the beta cal so the pointer sits over the furthest right hand lines on the meter scales. Don't touch that again for the short time you are using it at one sitting. These meters do not have the best balance, so set the cal with the meter level and do all your testing in the same position. Also, try to keep the transistor you are testing at room temperature. Don't hold it in your hand or fingers.

Hmm, that is how I thought the scale switch should work, so my confusion comes from the fact that it's apparently not working correctly. If it were working correctly, a reading of 70 when set to X1 should read 7 when set to X10, right? That's not what's happening now.

It does seem to calibrate correctly using the resistor and following the directions (following directions is the one thing I can do competently!), so I'm wondering if maybe one or more switches is giving intermittent contact. I measured voltage on the battery on the leads soldered to the case, so the battery is making good contact and supplying the correct voltage to the circuit.

Is the gain reading on my DVM accurate enough to use for matching?

Last edited by fredlf; 9th February 2010 at 05:20 PM.
  Reply With Quote
Old 9th February 2010, 07:31 PM   #2
anatech is offline anatech  Canada
diyAudio Moderator
 
anatech's Avatar
 
Join Date: Jun 2004
Location: Georgetown, On
Hi Fred,
You're ahead of me then. The fact that you measured your battery voltage in the circuit, rather than at the battery terminals really shows me that you like to confirm the results you are looking for. An excellent trait, one I have seen in few technicians. There's an "atta-boy" for you. No kidding, I am impressed.

Now that you are making me think about how these meters actually work, I had to actually check the switch using a transistor. The fact is, it's been over 25 years since I have even thought about how this thing is supposed to work - to put it into words. It's something I use subconsciously without even thinking about. So, if you are on the X10 scale and read a beta of ~9 (X10=90), throwing the switch to X1 should cause the pointer to move downscale to where you would imagine a reading of 90 would be. Reading a beta over 100 is really a very rough guess, and a reasonable setup since most transistors tended to have beta readings of 100 or less. Our newer signal transistors were not something they figured you would need an accurate answer on anyway. That's because temperature plays such a big role in these readings. For the same reason, your DVM reading is only a very approximate thing, the danger here is that it gives you actual numbers. That fools you into thinking you have more accuracy that you actually do.

For matching power transistors, use the IT-18 while being careful not to handle the transistors. Room temperature is the best you can do. Once things start heating up, any hope of a match is greatly reduced. If you are matching signal transistors, you pretty much have to use the jig I mentioned. At least both transistors are at a similar and are running the same current if a match is indicated. I do use an IT-121 for more accurate or detailed measurements (or to test J-Fets), but it takes too long to use in bench work. I actually got that as an unbuilt kit about 8 years ago from my father. It was fantastic to be able to build a Heathkit after so long. I think we have missed out on a lot since Heathkit stopped selling kits.

When running a service shop, I tried to get the guys to use things like the IT-18 for the simple reason that it re-enforced that the readings were approximate. One thing you will find is that if you have a couple transistors that measure very close in gain, the pair that have the same cal point on the meter will match pretty closely. Earlier, I think I told you to leave the calibration control the same for all measurements, that was wrong - sorry. I was doing something else (for which this was true) and advised you incorrectly. You calibrate for each transistor. If you warm the part up, or cool it, the calibration point will drift. If you have just de-soldered a transistor, you have to let it cool before trying to take an accurate reading. But keep in mind that for general circuit operation, having a gain that is "in the correct range" for that type, with no leakage, is what you are looking for. Abnormal high or low readings are cause for concern. I retest those and call them bad if they test the same.

I hope that helps Fred, Chris
__________________
"Just because you can, doesn't mean you should" my Wife
  Reply With Quote
Old 10th February 2010, 03:23 AM   #3
fredlf is offline fredlf  United States
diyAudio Member
 
Join Date: Dec 2009
Location: Portland, OR
Send a message via AIM to fredlf
Okay, so if I understand you correctly, I was wrong before. My DVM and the IT-18 ARE using the same scale. So I can use the DVM as a "reality check" against the IT-18, right?

Doing that with a 2SC2912, I get 264 stable on the DVM, but 30 on the IT-18 (X1 scale) or over 100 (X10). It does calibrate normally, meaning I can hook up the transistor and use the beta-cal control to move the needle to the CAL mark.

One thing odd about this IT-18 (an ebay purchase) is that instead of the little 10 kohm leakage adjust pot on the back of the function switch, the builder wired up a 10 kohm shafted pot on leads. I assume they did this in order to the calibration procedure without having to pull the chassis out of the box. I don't see how this would make any difference electronically, but I don't know much.

Anyway, I don't think this IT-18 is working right, at least not the beta testing amp. The leakage tests appear to work, but I don't have any way of proving that (other than shorting the lead).

All this transistor testing chat should be in a different thread. Chris, can you move it as a moderator?
  Reply With Quote
Old 10th February 2010, 05:17 AM   #4
anatech is offline anatech  Canada
diyAudio Moderator
 
anatech's Avatar
 
Join Date: Jun 2004
Location: Georgetown, On
Hi Fred,
Your wish is my command ... posts split and moved.

If you are getting numbers that are close between them, yes. Each will provide a sanity check for the other. A "cross check" in lab speak. The fact that you can calibrate normally is a very good sign.

Quote:
the builder wired up a 10 kohm shafted pot on leads.
Oh darn! I hate it when people get creative for no good reason. There is no good reason to make this control accessible at all. Set and forget is what this thing is all about. I was thinking, "how can you mess up a simple meter?" Looks like someone found a way all right!

Quote:
I don't see how this would make any difference electronically, but I don't know much.
One should never deviate from a working plan without a very good reason. In this case, the pot may have the wrong taper, the leads may cause trouble due to inductance or noise pickup. My advice would be to locate a control that can be mounted as close to the stock position as possible. Once you have the control, carefully check all the wiring and install the new control in the correct location. Then, run through the calibration procedure, it's easy as you know. It might be worthwhile to use a solder sucker to remove the bulk of the solder off the connections. Look for tarnish or black deposits. Remove any you find, then apply fresh solder. Not too much, just do it like the manual shows. This will eliminate any bad solder connections. Do not just add solder, that will not help you at all. In fact, you could end up with solder stuck inside components or shorting. When you use a solder sucker, it removes most of the heat energy with the molten solder. Then you can inspect the surfaces of the conductors and re-solder everything properly.

While I'm thinking of it, watch out for pencil marks or fluid residue marks. These can be conductive

You can test leakage with your finger or a high value resistor. Shorting the leads is hard on the meter.

Once you have it running, it will last you years. After all, I'm still using the one I bought as a kid, and it's been in the hands of idiots at my shop. I was dropped twice, breaking the meter cover twice. It's glued together and looks like heck - but it works fine.

-Chris
__________________
"Just because you can, doesn't mean you should" my Wife
  Reply With Quote
Old 10th February 2010, 09:44 PM   #5
fredlf is offline fredlf  United States
diyAudio Member
 
Join Date: Dec 2009
Location: Portland, OR
Send a message via AIM to fredlf
Well, I guess the ideal thing to do would be to rebuild it from the beginning. Then I can be positive it's wired correctly and I can test the values of the resistors and cap.

But I'll start by checking the switches for proper continuity and checking their wiring to see if it's correct.

Can you tell me more specifically how you would test the leakage functions with your finger? Do you mean using it as a resistor of sorts?

I can tell it's worth putting some time and effort into this thing, I just wish I had more of both of those to spare
  Reply With Quote
Old 10th February 2010, 10:00 PM   #6
fredlf is offline fredlf  United States
diyAudio Member
 
Join Date: Dec 2009
Location: Portland, OR
Send a message via AIM to fredlf
Quote:
Originally Posted by anatech View Post
My advice would be to locate a control that can be mounted as close to the stock position as possible.
How about this: M64Y103KB40 Vishay/Spectrol Trimmer Resistors - Multi Turn

I might have to use a quarter-inch or so fly-out to extend the leads to the width of the holes/lugs on the rotary switch. I could use shielded but it's hard to imagine the circuit is that sensitive to inductance. (Not to mention that would be some dang fiddly soldering.)
  Reply With Quote
Old 10th February 2010, 10:11 PM   #7
diyAudio Member
 
Join Date: Nov 2002
Quote:

There's an "atta-boy" for you. No kidding, I am impressed.


You don't see that everyday
  Reply With Quote
Old 10th March 2010, 09:13 PM   #8
fredlf is offline fredlf  United States
diyAudio Member
 
Join Date: Dec 2009
Location: Portland, OR
Send a message via AIM to fredlf
So the wiring appears to be correct (matches the manual). I checked all the resistors and they are the correct values. I checked the x10 switch for continuity and it is working correctly. I cleaned everything again with deoxit contact cleaner.

But still, I don't think the gain measuring circuit is working correctly. I can do the "new battery" calibration correctly. And, I can adjust bias cal correctly (slowly moving the needle to the cal point when a transistor is connected), but pushing the test switch always deflects the needle to around 60, no matter the transistor. In other words, a transistor that reads 200 hfe on my DVM still reads around 60 on the IT-18. The reading does not change significantly, or logically when flipping to the x10 range. In other words, the needle only deflects a little, if at all, so that 60 becomes 500 (50x10).

I didn't test the capacitor yet, and I'm still using the shafted pot. But I don't see how those would cause the behavior I'm seeing, given that I can calibrate normally using the resistor procedure, and that biasCAL works when a transistor is connected.

In other words, stumped again.

Last edited by fredlf; 10th March 2010 at 09:17 PM.
  Reply With Quote
Old 12th March 2010, 03:52 AM   #9
PB2 is offline PB2  United States
diyAudio Member
 
PB2's Avatar
 
Join Date: Sep 2004
Location: North East
Blog Entries: 1
I bought one of these on ebay recently and the X1/X10 switch was put in upside down so that they were opposite to the labels.
__________________
Pete Basel
http://www.linkedin.com/in/petebasel
  Reply With Quote
Old 12th March 2010, 09:03 PM   #10
anatech is offline anatech  Canada
diyAudio Moderator
 
anatech's Avatar
 
Join Date: Jun 2004
Location: Georgetown, On
Hi Pete,
Takes all kinds of people I guess! Not too proud of his workmanship either. The fault was never corrected.

Hi Fred,
I have a picture of how mine looks. It was factory wired (it was the only one they had - I wanted the kit). May have been built at the store, I can't say for sure. I was told it was new, and it appeared to be. I've included some scans of the manual, I hope they can help other members who are looking in, trying to help. I think you have the manual, right?

When you set the meter pointer on the "Cal" mark in preparation to measure beta, you are actually setting the emitter current to 1 mA - almost exactly. This information can be useful, and you can verify the current by putting the emitter lead in series with your multimeter (switched to current mode) and read the value. That will provide a great test to see how close you are to the way it should be working. There is also the cal procedure using a 180 ohm resistor. I bought a 1% resistor (actually a bunch, and selected the most accurate one) and fixed it inside. That allows me to have a good idea if anything is changing with the meter. This brings up an important point. This meter is sensitive to the voltage across the battery. Since I use it a lot, I've been thinking of building a power supply for it. This meter does not go through batteries, but I'd like to eliminate the variable created by an aging battery.

This meter is about the best there is for service bench use. It tells you exactly what you need to know, and does it quickly. There are more accurate meters out there, and Heathkit made some of those too. But they take more time to actually take a reading. This darn meter is pretty simple, and yet manages to be very accurate when compared to testing the transistor in a circuit where you can measure the actual currents and calculate the beta. If there is one meter I wish was available new, this is it.

I did buy a Heathkit IT-30 (?) [See below] transistor checker, but it went missing in the mail and I never did get it. It had a circular dial scale on the front panel and I wanted to restore it and see how close it was. Does anyone have one of these they don't want? I'll admit here and now that I have a weakness for test equipment, and I seem to have collected a number of Heathkit stuff.

I checked on Google, the transistor tester I didn't get was an IM-30. Here is a link that shows what it looks like. It seems to be more complicated to set up for service use than the IT-18. I really can't think of a better meter to use.

Why do I think so much of the Heathkit IT-18? Well, aside from being quick to use, it provides very close readings for beta. It also reminds the user that the reading is approximate due to the analog scale. A digital reading is useless and implies accuracy that does not exist. Also, this meter will read collector - base leakage current, and collector - emitter leakage current. except for shorted transistors that I find with my DVM, the leakage tests catch more failure modes than any other test I can think of. There are a small percentage that the beta test catches as well. Any transistor that has abnormally high or low beta for the type is condemned in my world.

I highly recommend this meter for bench use.

-Chris
Attached Images
File Type: jpg IT18-1.jpg (97.7 KB, 271 views)
File Type: jpg IT18-2.jpg (108.7 KB, 251 views)
File Type: jpg IT18-3.jpg (84.6 KB, 250 views)
__________________
"Just because you can, doesn't mean you should" my Wife
  Reply With Quote

Reply


Hide this!Advertise here!
Thread Tools Search this Thread
Search this Thread:

Advanced Search

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are Off
Pingbacks are Off
Refbacks are Off


Similar Threads
Thread Thread Starter Forum Replies Last Post
Looking for a manual for a Heathkit IT-18 fredlf Equipment & Tools 9 26th January 2012 07:05 PM
Transistor Tester Current Levels Stocker Parts 87 22nd December 2010 03:30 AM
Best transistor substitutes for Heathkit ar-15 ar15guy Solid State 4 8th April 2008 02:21 PM
HOW to use a Heathkit Model 10-18 oscilloscope Dusk Solid State 7 14th October 2004 10:00 AM
transistor tester janey Solid State 0 4th January 2002 02:44 PM


New To Site? Need Help?

All times are GMT. The time now is 06:51 AM.


vBulletin Optimisation provided by vB Optimise (Pro) - vBulletin Mods & Addons Copyright © 2014 DragonByte Technologies Ltd.
Copyright 1999-2014 diyAudio

Content Relevant URLs by vBSEO 3.3.2