Noisy output of Focusrite Scarlett 2i2 2gen

Hi,
I've been using the Focusrite Fcarlett 2i2 2gen interface for measurements for some time now.
The card seems to measure well. All the results from RMAA are at a similar level to the results I found online. I am posting a THD loop-measurement made with the RMAA.

I recently connected the output of the card to an oscilloscope (one of the differential outputs or the output from the differential to single ended shorting adapter - no difference).
To my surprise I saw a very noisy waveform. I'm posting screenshots from the oscilloscope for a 1kHZ -0.5dBFs waveform and for the output with no signal. As you can see, the noise reaches 200 mVp-p. I also include FFT spectra of these signals, where you can see that the noise spectrum is shifted to 30-40kHz which explains the good measurements in RMAA.

Is this normal for this interface?

I took measurements with the card connected to a laptop powered from the mains and battery - no differences

I will add that the output signal from other sources I own is clean in the same measurement setup .

Could someone confirm these results on their unit?

Marek
 

Attachments

  • THD_test_loop.png
    THD_test_loop.png
    12 KB · Views: 96
  • output_1KHz_-0.5dBFs.png
    output_1KHz_-0.5dBFs.png
    20.3 KB · Views: 100
  • output with no signal.png
    output with no signal.png
    44.8 KB · Views: 101
None of the above, this is how noise-shaping DACs work, they push the quantizatiobn noise up into the ultrasonic region where it can be filtered out or simply ignored. In fact they generate more noise that other types, but less noise in the audio band. The noise spectrum is "shaped" by clever high-order delta-sigma modulation schemes.
 
  • Like
Reactions: 1 user
None of the above, this is how noise-shaping DACs work, they push the quantizatiobn noise up into the ultrasonic region where it can be filtered out or simply ignored. In fact they generate more noise that other types, but less noise in the audio band. The noise spectrum is "shaped" by clever high-order delta-sigma modulation schemes.
That makes sense. This can also be seen on fft spectra. So the type of converter plays a role and this behavior is normal....

However, just to be sure, could someone take a screenshot of the waveforms from their scarlet 2i2 to compare?
 
diyAudio Chief Moderator
Joined 2002
Paid Member
I don't have a picture but I have seen a friend's Scarlet doing like that on a scope. That's what I meant with ultrasonic noise of the interface, the quantization noise. I have an older other brand interface which shows up even worse on a scope. If there's option to set a low pass digital filter in the tens of kHz region in the math menu of your scope you will see the noisy line getting thinner.