Fluke 77 service

Status
This old topic is closed. If you want to reopen this topic, contact a moderator using the "Report Post" button.
been thinking about this for some years now. and i´ll post it here, who knows maybe somebody will answer?.

problem:
got this Fluke 77 in early/mid 90´s and been happily using it ever since. but somwhere along the way it has stopped reading Mohms properly.
let´s say i take 5Mohm resistor an meashure it. it starts at 5M and counting down. what could be wrong?
i have never tinkered with such a device before. only hifi.
not really much of a problem but i would like to dedicate my better half some attention.

solution:
maybe i should do some service to it. are there any multimeter servicemen here willing to share some knowlage?

regards Fredrik
 
hi!

thank you!

i downloaded some manuals and will try to meashure the voltage as you described this weekend. i´ll report back here with results.

the battery is fine, should have mentioned that in the first post.

the meter i have is a 77 non MK, old but i like it.

the problem i described only occur in the MOhm range

it´s been like this for a very long time.

not that i do any meashuring in that range. but its nice to know your stuff is working correctly.
 
Administrator
Joined 2004
Paid Member
Hi figge 77,
I used to repair Fluke meters, and calibrate them also. Possibly some of these suggestions may help you.

Clean the banana jacks inside using alcohol only. Cotton swabs work, remove some cotton if the fit is too tight.

Clean the function switch metallic contacts. Alcohol again.

Check and make sure the PCB of you meter is clean. Alcohol would be the cleaner of chose here (like there is any question). Do not clean the entire PCB, or anywhere near the display contacts or IC unless nothing else works. If your display becomes faulty, clean the elastomer gently on both ends and also the PCB contacts and display contacts. The contacts are hard to see as they are transparent. Do not allow your finger oils or anything else to contaminate any surfaces in there.

If none of that works, I would normally resolder the terminals. I think that one was metal cylinders only. This is a job for someone who is very good at soldering and "has the touch".

Meters like the 87 types used a plastic terminal assembly that you just replaced. It used 4 wire connections to the common and volt terminals. Quite advanced for a hand held meter.

-Chris
 
thanks chris!

i resoldered the terminals and when i was at it all the other joints aswell. cleaned it up best i could. but still same thing.

it seems i lied in the posts before. the problem is not just at M range. i tried a 660k resistor and it showed 600k.
so i suspect the problem starts around 0,5Mohm or so. but at the higher values it cant get a reading it just falls .
 
Administrator
Joined 2004
Paid Member
Hi figge 77,
It still sounds like leakage in the range switch or PCB. The 77 is a pretty good meter and might be worth having Fluke look at it. Otherwise, buy a new Fluke.

I had been lucky in that I could fix most of them properly. Usually the only ones I gave up on were things like measuring the current across a 550 VAC circuit. Tends to vaporize the traces and severely mess with everything else. That and when they get squished flat by construction equipment.

-Chris
 
I don't mean to hi-jack the thread but I figured I'd ask a question since it was relevant.

I spent my graduation money on a Fluke 87 some 22 yrs ago and it still works flawlessly EXCEPT for the screen readout.. It started to fade a few years ago. I have cleaned the screen and PCB contacts and that generally helps but doesn't cure it completely.

How do I clean the pink "pads" or connector cables that tie the two together? Or should I just get replacements?

My 22 yr old fluke, my 21 yr old Ungar heatshrink gun and my 35 yr old Ungar fixed temp soldering station are all still rocking along. They just don't build them like that anymore... I sure hope I last as long as they do!! :D :clown:
 
Administrator
Joined 2004
Paid Member
Hi Jam,
An eraser is destructive to the elastomer. So is any form of cleaning, so it must be done gently and using a damp cotton swab. Isopropyl alcohol only (the 99% stuff). It can be purchased at any pharmacy or ordered in. It is not expensive.

Hi troystg,
To clean your display, there are four surfaces for each elastomer to work on. Do both sides of the elastomer ** gently ** using a damp cotton swab only. Then, clean both the contacts on the display and the PCB carefully. Don't be cheap, use a new swab. Be very careful not to allow anything to contact the cleaned edges. Reassemble them as soon as they are dry, don't leave them hanging around while you work on something else. This is also a good time to clean your lead terminals using the same materials. Never force anything into those terminals! Keep cleaning until the swab comes out clean (like cleaning a gun barrel). Again, don't use too much fluid or you will force the materials inside the terminals out onto the PCB. You do not want to do that.

As Jam correctly pointed out, buy new ones if that didn't work. Oddly enough, I did run into one where the contacts had somehow been worn off the glass display, or maybe had some lacquer or enamel on the contacts. It's well worth the cost of a new display if you need one. Your calibration should be close, even though you are well outside the calibration interval (by about 11 times in commercial use). :D Got your money's worth, didn't you?

Nice meter by the way. I am hoping to repair a junker to I can have one too. The new meters are "closed case calibration", which is really nice. It really helps the high frequency adjustment, which changes even by putting a different top cover on. I had to make a jig in order to "optimize" the calibration on these. It works extremely well.

-Chris
 
Hi Chris,

Oddly enough the eraser trick was told to me by a tech at Fluke years ago, it worked on my co-workers Fluke though you had to be very careful and ran the risk of damaging the pads.
I prefer the alcohol method myself and you are right as I should have qualified it by saying Isopropyl alcohol.

Jam
 
Administrator
Joined 2004
Paid Member
Hi Jam,
I can't fault you for not specifying the type of alcohol used. This is the first time in this thread that I have as well. :)

The eraser on an elastomer idea makes me grit my teeth just thinking about it! The possibility of damaging the elastomer is very high.

Erasers used on PCB contacts is also a very bad practice. Try that in Agilent's service center and watch what happens! They have proved this technique will probably cause damage as well. You need high magnification to see the damage start, but that fact is undeniable.

Best, Chris
 
according to the servicemanual:

if ohms reads low or won´t read ol----> shorted Q1 and Q2 or leaky E1 or E2

the E1 and E2 are 1500V surge protectors. can i test them if they are ok somehow? or what can i use as replacement? a wire bridge hehe?

just kidding

will check the transistors aswell but aint gonna be inte workshop in a couple of days.
 
Administrator
Joined 2004
Paid Member
Hi Troy,
Hmmm, Will have to write the moderators..
:eek:

I R 1

Now I feel completely forgotten, ignored. Just another old, forgotten tech from yesteryear.

Hi figge77,
Well yes, check the input circuit. This is the time when a good bench meter comes in handy. You can simply remove them for testing. I have seen a number that were either zapped or had some conductive residue in there.

They should test as an open, however, if you remove them and start getting better readings - guess what?

-Chris :)
 
Administrator
Joined 2004
Paid Member
Hi Troy,
:D
What can I say?

Hi Jack,
Now, if you just nailed the fault, I'll be laughing pretty hard. Still, there is a "low batt" indicator on the display.

Now that you mention it, I do recall having to replace battery clips often enough on these meters. My memory needed a jog too - thank you!

It's generally the simplest faults that mess us all up.

Hi figge77,
Do you have the manual or schematic for that meter? This fault shouldn't really be a terrible thing to troubleshoot.

-Chris
 