I have an old DANA 4600 bench top DMM that I set out to 'calibrate'. Well, really just compare the readings to a new, more accurate meter and tweak it if required. I also replaced the electrolytic caps in the power supply section just because I have some other equipment of this same generation that had dried out caps. When I went to check things out, I find that the power transformer secondaries are putting out too high a voltage.
The transformer has three secondaries:
16V reads 19V (19% high)
36.5V reads 42.2V (16% high)
15.3V reads 17.4V (14% high) - this one isn't loaded, it is for a data option the meter doesn't have.
I measured the line input at 122V. The transformer is the 'universal' design that allows it to be configured for 100V, 120V, 220V and 240V operation by changing the primary wiring. I verified that it is set for 120V operation.
The 16V winding is center tapped and then full wave rectified. The DC voltage should be 9.0 - 9.1V, but reads 11.1V (23% high). The filter cap is only rated at 10V, so it definitely shouldn't be this high.
The only thing I can think of that would cause the secondary voltage to be too high is a partially shorted primary. The transformer doesn't get hot. Barely warm. If the primary was partially shorted, wouldn't it be running hot? I've never seen a transformer fail like this. Does it ever happen?
I need to do something about it. The 5V regulator (a simple zener biased emitter follower) runs too hot. The zener gets too hot after about half an hour of operation and the voltage starts to rise. The 5V output gets as high as 5.6V (not good since it supplies TTL chips). The emitter follower transistor is also way too hot to hold your finger on.
Terry
The transformer has three secondaries:
16V reads 19V (19% high)
36.5V reads 42.2V (16% high)
15.3V reads 17.4V (14% high) - this one isn't loaded, it is for a data option the meter doesn't have.
I measured the line input at 122V. The transformer is the 'universal' design that allows it to be configured for 100V, 120V, 220V and 240V operation by changing the primary wiring. I verified that it is set for 120V operation.
The 16V winding is center tapped and then full wave rectified. The DC voltage should be 9.0 - 9.1V, but reads 11.1V (23% high). The filter cap is only rated at 10V, so it definitely shouldn't be this high.
The only thing I can think of that would cause the secondary voltage to be too high is a partially shorted primary. The transformer doesn't get hot. Barely warm. If the primary was partially shorted, wouldn't it be running hot? I've never seen a transformer fail like this. Does it ever happen?
I need to do something about it. The 5V regulator (a simple zener biased emitter follower) runs too hot. The zener gets too hot after about half an hour of operation and the voltage starts to rise. The 5V output gets as high as 5.6V (not good since it supplies TTL chips). The emitter follower transistor is also way too hot to hold your finger on.
Terry