Old 1st April 2009, 01:37 AM   #1
diyAudio Member
 
Join Date: Jun 2008
Default B&K 500 Tube Tester Calibration

I could use some help from someone who has a copy of the calibration procedure for a B&K 500. Specifically, I need the resistor value used to set the sensitivity. The 707 uses a 6K and the 650 uses a 2K, so I don't think I can just wing it here.

I'll also need to know where (what socket and pin) to apply that resistor.

Also, I'd like to know where to apply the 2meg resistor for the "shorts pot" adjustment.

Thanks

Dave
  Reply With Quote
Old 1st April 2009, 04:22 PM   #2
diyAudio Member
 
Join Date: Jun 2008
Woops, I said sensitivity; I ment ZERO.

The resistor is used to set the ZERO.

Any help would be appreciated.
  Reply With Quote
Old 1st April 2009, 09:02 PM   #3
Cobra2 is offline Cobra2  Norway
DIY !
diyAudio Member
 
Cobra2's Avatar
 
Join Date: Apr 2002
Location: Stavanger, Norway
Send a message via Skype™ to Cobra2
David Campbells instructions:
-------------
Signal Voltage adjustment (Test Switch in Gm position)
Set to 1.0 volt AC at terminal 5 of the wiring block on the top section of sockets to terminal 1 (gnd), Terminal 1 is most near socket 7. The terminals are numbered 1 through 12 on the schematic. This AC voltage is adjusted by R-14 which has yellow, black and orange wires attached, and mounted left of another Potentiometer when the meter is facing you under the top section of sockets mounted to the bottom of the tester. Terminal 5 is the low bias -2.5VDC voltage terminal, i.e. at that terminal you have both a DC bias and an AC signal voltage. See the following pictures for Terminal and Pot locations.
-----------------
DC BIAS Adjustment (Test Switch in Gm position)
Proper DC bias at terminals 5, 6, 7 to terminal 1 (gnd) is accomplished by adjusting the R-9 potentiometer which is positioned just right of R-14. Set a voltmeter to DC volts; connect between terminals 5, to gnd (set this voltage to -2.5VDC, terminal 6 and 7 voltage adjustment is done when terminal 5 is set.

Terminal 5 is -2.5volts, Terminal 6 is -7.5 volts, Terminal 7 is -13.0 volts
-------------------------
Balance Control Adjustment (Test Switch in Gm position)
Connect a 6k ohm 10 watt resistor between terminal 10 (plate) and terminal 1 (gnd) with Sensitivity Control set to 100. Adjust R-19 so that you get zero reading on the tube tester meter. R19 access hole is left of the 6AT6 tube under the top section of sockets.
------------------------
Short Sensitivity (Test Switch in Short position)
I donít see any adjustment for this (No more Pots). You can check to see if itís working by inserting a 1 Meg ohm resistor between pins 2 and 4 of socket 7A and looking at the shorts lamp. This lamp must be a NE-51 for the circuit to work properly.
------------------------
Gas Sensitivity (Test Switch in Gas position)
Again, I donít see any adjustment for this (No more Pots again). You can check this by inserting a 20 Meg ohm resistor between pins 1 and 2 of socket 7A and seeing the meter move to about 2000 on the black Gm scale. On my tester, the reading was about 1800.
------------------------
Filament Voltage Check
I checked each of the voltages for the filaments with no tubes installed first. They were each slightly higher then the settings. I then checked the 6v setting with a 5881 tube which takes about .9 amps of heater current. I measured 6.1 volts, which indicates that the transformer is working well.
--------------------

Arne K
__________________
Don't believe everything you think...
  Reply With Quote
Old 3rd June 2010, 07:18 AM   #4
diyAudio Member
 
Join Date: Jun 2008
Arne,

I just noticed that I never thanked you for your assistance last year. My bad; your assistance was very helpful. Thanks. I have three comments to add to the procedure above.

On my tester, there is a plastic acorn nut just above the power toggle switch. Some models have a jewel power lamp in that location instead. Removing the acorn nut reveals a pot for adjusting the gas.

If you lift the lower panel and look under the shorts lamp, you will find another pot that is the adjustment for shorts. I use a 2 meg resistor to gain a little more sensitivity to shorts.

The balance can be tested without lifting the panel by placing the 6K resistor between pins 2 and 5 of socket 7A. This technique is lifted from the 707 procedure.

Dave

Last edited by Captn Dave; 3rd June 2010 at 07:30 AM.
  Reply With Quote
Old 4th May 2011, 05:27 AM   #5
diyAudio Member
 
Join Date: Aug 2010
Quote:
Originally Posted by Captn Dave View Post
If you lift the lower panel and look under the shorts lamp, you will find another pot that is the adjustment for shorts. I use a 2 meg resistor to gain a little more sensitivity to shorts.
I don't see it. I see R19... for the balance adjustment. Or am I wrong? Are you seeing what I see? That pot definitely adjusts the balance on my tester. Maybe I'm wrong... Can you post a picture?
I just got a BK-500 off ebay. Not in bad shape... But I can't get anything to not test as shorted. Most of the tubes I tried to test took a few seconds or more to slowly light up the neon bulb. One of the 6L6s I tried lit up bright immediately. I think that one was bad... pretty sure the the others are good. I just think the tester is out of spec and needs to be adjusted. Has anyone found any other info on this?
If I don't get any replies soon then I will start changing caps and post the results. Could be a resistor that's drifted out of spec. No reason not to continue compiling info on these testers. Anyone who is willing to work on one of these old testers surely deserves to have it work as good a possible. As old as they are, the info should be public domain by now.

Tony
  Reply With Quote

Reply


Hide this!Advertise here!
Thread Tools Search this Thread
Search this Thread:

Advanced Search

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are Off
Pingbacks are Off
Refbacks are Off


Similar Threads
Thread Thread Starter Forum Replies Last Post
Calibration of Mercury 1000 tube test set TheGimp Tubes / Valves 15 8th March 2011 04:55 AM
Hickok tube tester calibration help needed tc4all Tubes / Valves 4 12th January 2009 03:48 AM
Tube Tester jtwrace Tubes / Valves 5 8th September 2007 05:47 PM
Tube tester Sjef Tubes / Valves 6 29th January 2007 09:38 PM
help using Tube Tester ebarker2 Tubes / Valves 3 31st January 2004 03:30 AM


New To Site? Need Help?

All times are GMT. The time now is 10:16 AM.


vBulletin Optimisation provided by vB Optimise (Pro) - vBulletin Mods & Addons Copyright © 2014 DragonByte Technologies Ltd.
Copyright ©1999-2014 diyAudio

Content Relevant URLs by vBSEO 3.3.2