Old 1st January 2013, 05:11 AM   #1
chanwoo is offline chanwoo  South Korea
diyAudio Member
 
Join Date: Mar 2006
Location: Seoul
Default Dark CRT of Tektronix 576 curve tracer

Dear. Friends

I recently bought defected Tektronix 576 curvetracer to pass time.

I found step generator is not oscillating because rusty ICs pin. Replaced all of them. However major trouble is low intensity of CRT. I checked HV for CRT and actual voltage was 3.2KVDC but it suppose to be 4KVDC. And 6.3VAC heater voltage and it's lit fine.

When I turning intensity contron from low to high the spot or line on CRT the intelsity increasing was not linear. The intensity is changing from dark, low, mid low, high and mid low again on it's way. Therefore I have to negotiate with 60% of intensity control position to get maximum intensity but the intensity is still not enough. Really hard to see line on CRT when the spot on CRT is stratching as a line.

Focus control works fine because spot is changing oval shape from horizontal to vertical.

Here is my question, Is HV3.2KVDC not enough for this CRT to get enough intensity or the CRT immission too low?

It is very hard to find replacement of the CRT. Has anyone solution or apare part for sale?

Thank you in advanced.
  Reply With Quote
Old 1st January 2013, 05:26 AM   #2
freax is offline freax  Australia
diyAudio Member
 
freax's Avatar
 
Join Date: Sep 2007
Location: NSW
Has the intensity pot been cleaned?

I would check that and whatever it connects to for dry joints.

Have you checked the B+ voltage and other power supply voltages? If you are getting low HV then its likely that the power supply feeding your flyback is lower too.
__________________
"It seems that perfection is reached not when there is nothing left to add, but when there is nothing left to take away" - Antoine de Saint Exupéry

Last edited by freax; 1st January 2013 at 05:44 AM.
  Reply With Quote
Old 2nd January 2013, 02:50 AM   #3
chanwoo is offline chanwoo  South Korea
diyAudio Member
 
Join Date: Mar 2006
Location: Seoul
Freax,

I checked the intensity pot using DVM and the pot works good. Also I found the spot light is scattering when I turn intensity pot to full clockwise. (dark, low, mid low, high and scattering) Which means spot is scattered when I turn intensity pot to full clockwise)

I have checked all power supply voltages and all they are within spec.

I don't know what reason the spot is scattering? I will attach photo for your reference later.

Thank you.
  Reply With Quote
Old 2nd January 2013, 01:05 PM   #4
chanwoo is offline chanwoo  South Korea
diyAudio Member
 
Join Date: Mar 2006
Location: Seoul
Here are videos for your reference.

Tektronix 576 curve tracer low intensity - YouTube
Intensity control is not linear when turning intensity pot.


IMG 1511 - YouTube
Multiple lines for another example. More lines on CRT then dark more and more.


IMG 1512 - YouTube
What is this? Vert 20uA, Horiz 2V and turning variable collector supply dial. Is this noise?
  Reply With Quote
Old 3rd January 2013, 07:30 AM   #5
diyAudio Member
 
jazbo8's Avatar
 
Join Date: Jan 2011
Location: In Transient
Try TekScopes : All about classic Tektronix CRT o'scopes lots of experts hang out there.

Good luck,
Jaz
  Reply With Quote
Old 8th January 2013, 06:55 PM   #6
diyAudio Member
 
Loudthud's Avatar
 
Join Date: Nov 2007
Location: Near Dallas Texas USA
What do you mean when you say "Scattering"? Do you mean the spot gets bigger or that it moves? What did you measure the high voltage with? Tek usually used a Tripplet 630 VOM. 20000 ohms pre volt, 6000V range means 120 Meg input resistance. I would suspect a high voltage problem.
  Reply With Quote
Old 9th January 2013, 04:32 PM   #7
diyAudio Member
 
Join Date: Dec 2010
Quote:
Originally Posted by chanwoo View Post
When I turning intensity contron from low to high the spot or line on CRT the intelsity increasing was not linear. The intensity is changing from dark, low, mid low, high and mid low again on it's way.
This is called "double peaking", and it means the CRT is near the end of its life

Some things to try:

1. CRT rejuvenation: CRT Restoration for the (Brave) Experimenter

2. See if anyone on the TekScopes mailing list has an extra CRT

3. Connect an external display

4. Wedge in an LCD module
  Reply With Quote

Reply


Hide this!Advertise here!
Thread Tools Search this Thread
Search this Thread:

Advanced Search

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are Off
Pingbacks are Off
Refbacks are Off


Similar Threads
Thread Thread Starter Forum Replies Last Post
Using The Tektronix 575 Curve Tracer Mark A. Gulbrandsen Solid State 29 21st May 2012 07:01 PM
curve tracer? maxpou Solid State 4 24th October 2009 05:29 PM
Tektronix 370 Curve Tracer - ease of use? PB2 Equipment & Tools 0 20th November 2007 04:12 PM
Virtual Tek 576/577 curve tracer in EWB V10 jackinnj Equipment & Tools 0 1st March 2007 01:41 PM
FS: tektronix 575 curve tracer SQ Kid Swap Meet 3 11th July 2004 04:22 PM


New To Site? Need Help?

All times are GMT. The time now is 07:59 AM.


vBulletin Optimisation provided by vB Optimise (Pro) - vBulletin Mods & Addons Copyright © 2014 DragonByte Technologies Ltd.
Copyright ©1999-2014 diyAudio

Content Relevant URLs by vBSEO 3.3.2