My Fluke 11 is giving erroneous AC & DC voltage readings on the high side by about 70%!! AC voltages such as 120 (taken at an outlet) are reading at 205v, and DC voltages that should be 8.4 volts are measuring at 14.50 (70% over)!

I called Fluke and they said the meter is shot and suggested a replacement with a core exchange. Needless to say it was too expensive, and I would rather try to repair it myself if it is repairable. I downloaded the service manual and hopefully with this forum's help, I can repair it.

Any ideas as to what could be wrong? What should I be checking? Please let me know.
__________________
Samer
  Reply With Quote
Old 20th November 2011, 07:00 AM   #2
Francec is offline Francec  Australia
Banned
 
Join Date: Jul 2006
Location: Port Melbourne, Australia
As I understand it, Flukes are, largely, based on very big ICs so the advice may well be correct. I would like to be proven wrong being an owner of a Fluke DMM but I doubt it.
  Reply With Quote
Old 20th November 2011, 08:24 AM   #3
AJT is offline AJT  Philippines
diyAudio Moderator
 
AJT's Avatar
 
Join Date: May 2003
Location: Palatiw, Pasig City
did you replace the battery yet?
__________________
the best advertisement for a good audio design is the number of diy'ers wanting to build it after all the years....never the say so of so called gurus....
  Reply With Quote
Old 20th November 2011, 08:49 AM   #4
diyAudio Member
 
KatieandDad's Avatar
 
Join Date: Nov 2011
Location: UK
It's a long shot but try cleaning the switch contacts.
  Reply With Quote
Old 20th November 2011, 09:01 AM   #5
diyAudio Member
 
Join Date: Jan 2011
Location: Irvine, CA
Quote:
As I understand it, Flukes are, largely, based on very big ICs so the advice may well be correct. I would like to be proven wrong being an owner of a Fluke DMM but I doubt it.
And you would be right. The Service Manual lists two of them, but lists the P/N for only one of the two, so I assume it is the only one that is replaceable. I will open it up tomorrow and check.

Quote:
Originally Posted by Tony View Post
did you replace the battery yet?
I haven't, but as a matter of fact, the low battery indicator is on. Could the problem be so simple as to only require a battery change? I'll let you know if that solves it.
__________________
Samer
  Reply With Quote
Old 20th November 2011, 09:07 AM   #6
diyAudio Member
 
KatieandDad's Avatar
 
Join Date: Nov 2011
Location: UK
You can buy a new Fluke on E-Bay for less than the cost of the IC. As it will need calibrating afterwards its hardly worth repairing it.
  Reply With Quote
Old 20th November 2011, 09:24 AM   #7
diyAudio Member
 
Join Date: Jan 2011
Location: Irvine, CA
Default Problem Solved!

I couldn't go to sleep after I read the post about swapping the battery out (it's after 2:00 AM on the West Coast), so I changed it, and that took care of the problem! Thanks Tony!

How could the Fluke tech I spoke with on the phone not ask me about something like this? Their 1st and only answer was that the DMM had gone bad and needed to be replaced. Simply amazing!

BTW, neither of the ICs is serviceable/replaceable. They are big, and they are soldered to the board. So much for that idea.
__________________
Samer
  Reply With Quote
Old 20th November 2011, 09:26 AM   #8
diyAudio Member
 
KatieandDad's Avatar
 
Join Date: Nov 2011
Location: UK
SUCCESSFUL RESULT - That is the aim of the game.
  Reply With Quote
Old 20th November 2011, 10:36 AM   #9
AJT is offline AJT  Philippines
diyAudio Moderator
 
AJT's Avatar
 
Join Date: May 2003
Location: Palatiw, Pasig City
Quote:
I haven't, but as a matter of fact, the low battery indicator is on. Could the problem be so simple as to only require a battery change? I'll let you know if that solves it.
i have other dmm's that exhibit this symptoms when battery is low....battery is nominally 9 volts, at 7 volts these symptoms appear....

i also has this problem with some segments on the lcd not lit......solution is to clean the contacts with cotton buds dipped in rubbing alcohol, both pads and lcd contacts.....in case you encounter this problem....
__________________
the best advertisement for a good audio design is the number of diy'ers wanting to build it after all the years....never the say so of so called gurus....

Last edited by AJT; 20th November 2011 at 10:42 AM.
  Reply With Quote

Reply


Hide this!Advertise here!
Thread Tools Search this Thread
Search this Thread:

Advanced Search

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are Off
Pingbacks are Off
Refbacks are Off


Similar Threads
Thread Thread Starter Forum Replies Last Post
Fluke 8840A bench DMM for sale grenert Swap Meet 2 9th May 2011 04:26 AM
Fluke 175 Dmm ? semi932 Equipment & Tools 5 18th May 2010 01:25 AM
Get a Fluke DMM ThSpeakerDude88 Parts 107 21st August 2007 12:40 AM
FS: Fluke 87 DMM w/ case, lead, manual gengis Swap Meet 5 20th September 2006 05:09 PM
Fs Fluke 77 Dmm powertriode Swap Meet 2 22nd July 2006 07:52 AM


New To Site? Need Help?

All times are GMT. The time now is 07:14 PM.


vBulletin Optimisation provided by vB Optimise (Pro) - vBulletin Mods & Addons Copyright © 2014 DragonByte Technologies Ltd.
Copyright 1999-2014 diyAudio

Content Relevant URLs by vBSEO 3.3.2