Old 4th June 2010, 05:34 PM   #1
jblmar is offline jblmar  United States
diyAudio Member
 
Join Date: Dec 2005
Location: Brooklyn, New York
Default Sound Technology ST 1701A Not Settling In Distortion Mode

My 1701A works fine except that in Distortion mode, the pointer will not settle completly. It will waver +/- 2 or 3 lines. After selecting a frequency and reading the distortion on the meter, if I press the 100 Hz or X10 or X100 switches, the set reacts as if a new frequency was selected. This is in the LOW DISTORTION mode. The FAST mode LED lights (as it would changing frequency), the meter pegs to the right (as it should) then the LED goes off and the pointer settles. It doesn't fully settle. There are times when the pointer swings to 2/3 of the scale then back. Pressing the above switches a few times returns the set to somewhat normal operation.
Anyone know of a good contact cleaner to use?
  Reply With Quote
Old 5th June 2010, 03:54 AM   #2
1audio is offline 1audio  United States
diyAudio Member
 
Join Date: Mar 2004
Location: SF Bay Area
Blog Entries: 3
What you describe is correct behavior. When any change is made in the frequency or the switches are pressed it starts a timeout on the oscillator, part of the fast settling process.

How low does it go in self test 1 KHz 3V with the 400 and 30K filters on? If you get to .0015% you are near the performance limits. its possible to improve it from there but its a lot of work for a snall improvement.

ST warn's against cleaning the switches since the chemicals could cause leakage etc. I don't know what to say about a 20 year old switch array. If it was in LA you may need to. I would look at what Caig Labs DeoxIT® D-Series has that could be used.
__________________
Demian Martin
Product Design Services
  Reply With Quote
Old 5th June 2010, 09:23 PM   #3
jblmar is offline jblmar  United States
diyAudio Member
 
Join Date: Dec 2005
Location: Brooklyn, New York
Takes about 3 seconds to go back to LOW DISTORTION mode. Same as when it works properly.
I think I corrected the problem. The capacitors located on the Ratio switch or a bad ground connection most likely. Removed the top and bottom panel and the set works fine now.
I'll check and clean the ground connections.
Last resort would be to clean switches
  Reply With Quote
Old 5th June 2010, 09:29 PM   #4
diyAudio Member
 
john curl's Avatar
 
Join Date: Jul 2003
Location: berkeley ca
Just to contribute here, I have 3 ST 1700's in my lab at the moment, 2 are Audio1's and one is my own. They ALL have problems developed over decades. The switches stop working properly, the null circuits stop nulling, etc. Still, they are OK units, better than most stuff out there.
  Reply With Quote
Old 6th June 2010, 11:55 PM   #5
jblmar is offline jblmar  United States
diyAudio Member
 
Join Date: Dec 2005
Location: Brooklyn, New York
I suspect the Sprague Tantalum caps on the RATIO switch as being the cause. Connecting my DMM for voltage test to the caps shows the following when the meter swings wildly.
Withe the DMM connected to - of the caps, the side leading to the ratio switch, the voltage rises slightly as the meter swings. It remains stable oyjerwise.

With the + side of the cap connected (this side is directly coupled to the output of the chip), voltage remains stable in either state.
Just a first look, but I may be on the right track.

Anyone knows where I can get those Sprague caps?
  Reply With Quote
Old 7th June 2010, 04:13 AM   #6
1audio is offline 1audio  United States
diyAudio Member
 
Join Date: Mar 2004
Location: SF Bay Area
Blog Entries: 3
I don't have one handy to look at. Perhaps you can post a photo of the affected area. What options does it have?

The caps attached to the opamps at the output of the amplitude and phase null detectors are all large values. Are those what you are referring to? They are used for driving the photocells in the notch filters. They are used in integrating around the opamps that drive the leds connected to the photocells.

A good quality low leakage electrolytic would work OK. However I think you are looking at a symptom, not a cause. The network around the notch has the resistors from the tuning in it. An intermittent connection there would cause all kinds of grief. many of the connections in the box are using .025" square posts with push on berg connectors. These are quite reliable, for the first 10 years. Try remove/replace any around the switches to clean and renew the connections.

Do you have a scope to look at the unit?
__________________
Demian Martin
Product Design Services
  Reply With Quote
Old 7th June 2010, 02:11 PM   #7
jblmar is offline jblmar  United States
diyAudio Member
 
Join Date: Dec 2005
Location: Brooklyn, New York
Demian,
The caps are,
C313
C314
C315

C324
C325
C326

From 1 uF to 100 uF. 10%. 150 is printed on the cap.
I'll get some pic for you.
BTW. The distortion reading you asked about (post 2) is .001%.
All direct to the RATIO switches.
  Reply With Quote
Old 8th June 2010, 04:58 PM   #8
jblmar is offline jblmar  United States
diyAudio Member
 
Join Date: Dec 2005
Location: Brooklyn, New York
Quote:
Originally Posted by 1audio View Post
I don't have one handy to look at. Perhaps you can post a photo of the affected area. What options does it have?

The caps attached to the opamps at the output of the amplitude and phase null detectors are all large values. Are those what you are referring to? They are used for driving the photocells in the notch filters. They are used in integrating around the opamps that drive the leds connected to the photocells.

A good quality low leakage electrolytic would work OK. However I think you are looking at a symptom, not a cause. The network around the notch has the resistors from the tuning in it. An intermittent connection there would cause all kinds of grief. many of the connections in the box are using .025" square posts with push on berg connectors. These are quite reliable, for the first 10 years. Try remove/replace any around the switches to clean and renew the connections.

Do you have a scope to look at the unit?
Yes, I checked the oscillator output on the scope as well as frequency checks on my Fluke 1900A Counter. The pattern was excellent as well as the frequency on the 1900A. Rock solid over a few hour period.
The problem is either on the analyzer the switches or the caps.
  Reply With Quote
Old 8th June 2010, 06:47 PM   #9
1audio is offline 1audio  United States
diyAudio Member
 
Join Date: Mar 2004
Location: SF Bay Area
Blog Entries: 3
Let me recap before further exploration-
1) Oscillator is stable (What happens to it when you press a button?)
2) Analyzer intermittantly jumps around but will settle to .001% or so.

Please confirm the following:
1) pressing a button on the frequency array that is already selected doesn't produce a massive transient on the oscillator.
2) The range switch causes the instability on the analyzer- Going down range sometimes causes fluctuations.

Places to check- first are the optocouplers. They are soldered in (unless on later units they used sockets). See if they are microphonic or connections around them are. They may need replacement, they do age a lot. Check that they are in the proper operating range, look at the current through the LED (voltage on series resistor) to see that its right and stable.
The rotary switch may have dirty contacts. Its silver plated and Deoxit should work well on it.
Do you have the IM option or any of the other options?
__________________
Demian Martin
Product Design Services
  Reply With Quote
Old 11th June 2010, 05:26 PM   #10
jblmar is offline jblmar  United States
diyAudio Member
 
Join Date: Dec 2005
Location: Brooklyn, New York
emian,
Quote:
Originally Posted by 1audio View Post
Let me recap before further exploration-
1) Oscillator is stable (What happens to it when you press a button?)
2) Analyzer intermittantly jumps around but will settle to .001% or so.

Please confirm the following:
1) pressing a button on the frequency array that is already selected doesn't produce a massive transient on the oscillator.
2) The range switch causes the instability on the analyzer- Going down range sometimes causes fluctuations.

Places to check- first are the optocouplers. They are soldered in (unless on later units they used sockets). See if they are microphonic or connections around them are. They may need replacement, they do age a lot. Check that they are in the proper operating range, look at the current through the LED (voltage on series resistor) to see that its right and stable.
The rotary switch may have dirty contacts. Its silver plated and Deoxit should work well on it.
Do you have the IM option or any of the other options?
Demian,
To answer your questions.
Q1: When a preselected button is pressed (x10, 100 or x100 especially), the meter pegs in Distortion mode.
Q2: Never really settles and holds there. Always moving.

Q1A: Yes, produces 100% deflection of the meter in Distortion mode.
Q2A: Instability on all ranges that show a display.

I plan to view the output of the oscillator & analyzer sections on the scope over the weekend. I also plan to connect an ext oscillator to the 1701A and measure its distortion. This will indicate if the 1701A is reading the distortion of the 1701A oscillator or the problem is within the analyzer section.
Here's how to connect your 'scope to the 17xx;

From 17xx
Generator Out to Balanced Inputs.
Distortion Out to scope.
Monitor to scope.

I pulled and cleaned all Op Amps on both boards. Swaped like OP Amps to check if they were the issue. Also substituted a new NE5534H as a final check. Same results.
P.S. Mark and return OP Amps to original location.

When I first bought the 1701A (used) it arrived with little or no output from the oscillator as wll as a distorted pattern. Replaced U1 on the oscillator board. I had to by 10 pieces @ $8 each. Needless to say, I wasn't happy. On the up side, the oscillator now worked.

Cleaned INPUT, RATIO & ATTEN switches first with CRC then 100% Deoxit. ST warning relates to cleaners that were available 35 years ago.

I'll post my findings.

The 1701A has Option 003 (SET LEVEL).

Last edited by jblmar; 11th June 2010 at 05:37 PM.
  Reply With Quote

Reply


Hide this!Advertise here!
Thread Tools Search this Thread
Search this Thread:

Advanced Search

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are Off
Pingbacks are Off
Refbacks are Off


Similar Threads
Thread Thread Starter Forum Replies Last Post
Speaker cable for Definetive Technology ST jzvjg Everything Else 0 16th February 2010 04:58 PM
Any tips on ST-1700B Distortion Analyzer repairs..? vibroverbus Equipment & Tools 2 29th November 2009 10:35 PM
Sound Technology ST 1000A & 1701A jblmar Solid State 11 5th March 2007 02:31 PM
WTB: Full Manual For Sound Technology 1701A Mark A. Gulbrandsen Parts 2 23rd February 2007 03:36 PM
Sound Technology 1700B Distortion Measurement Systems for sale nwaero Swap Meet 0 15th December 2006 07:09 AM


New To Site? Need Help?

All times are GMT. The time now is 03:40 PM.


vBulletin Optimisation provided by vB Optimise (Pro) - vBulletin Mods & Addons Copyright © 2014 DragonByte Technologies Ltd.
Copyright ©1999-2014 diyAudio

Content Relevant URLs by vBSEO 3.3.2