Default Scope trace more and more to one side

Got a cheap scope that worked when it arrived, for about a week; it's a Tektronix T922R. All of a sudden, when I turned it on, the trace on both channels showed fine, but half of it outside the screen on the right side. The horizontal position button cannot bring it all the way to the center. As I turned it on and off a few more times, the trace moved more and more to the right, and now it's outside the screen.

Has any of you any ideas what might cause this?
  Reply With Quote
Old 14th August 2009, 04:30 PM   #2
syn08 is offline syn08  Canada
Account disabled at member's request
 
Join Date: Aug 2005
Location: Toronto
Default Re: Scope trace more and more to one side

Quote:
Originally posted by ikoflexer
Got a cheap scope that worked when it arrived, for about a week; it's a Tektronix T922R. All of a sudden, when I turned it on, the trace on both channels showed fine, but half of it outside the screen on the right side. The horizontal position button cannot bring it all the way to the center. As I turned it on and off a few more times, the trace moved more and more to the right, and now it's outside the screen.

Has any of you any ideas what might cause this?

http://www.logwell.com/tech/oscillos...921_922_sm.pdf

Ch. 7, "Sweep and Horiz amp"
  Reply With Quote
Old 14th August 2009, 05:27 PM   #3
iko is offline iko  Canada
diyAudio Moderator
 
iko's Avatar
 
Join Date: May 2008
Location: Toronto
Thank you, I've been browsing that chapter this morning. I was hoping someone would have had this problem, and have some specific advice, but I guess I'll have to start digging :/
  Reply With Quote
Old 14th August 2009, 06:18 PM   #4
diyAudio Member
 
HollowState's Avatar
 
Join Date: Aug 2006
Location: Taxland, New Jersey
First, verify that the horizontal mag control is fully counter clockwise in the X1 position. Next, you'll have to open the split case and check the low voltage power supply. This is the small board on the rear bottom side with the transformer. It has four outputs, +100v unregulated, +100v regulated, +8v and -8v. Measure these levels at the pin connector (P7) at the edge of the board. Pin 5 and the metal frame is the common ground.

Pin 1 is +100v reg.
Pin 2 is +100v unreg. (typically +120v)
Pin 3 is +8v
Pin 4 is -8v

While you are there, carefully inspect the solder connections at the P7 connector. Poor solder connections here are a very common problem in these scopes. This also applies to the other pin connectors used elsewhere. Both male and female counterparts.

If the output is low on one or more, you will need to find out why. Keep in mind that these supplies are current limited and will drop when loaded down heavily. So it could be the regulator itself or something downstream in the circuit.

If the supplies are up, you'll have to troubleshoot the horizontal section. (Remember those pin connectors.)
__________________
"I regard the brain as a computer which will stop working when it's components fail. There is no heaven or afterlife for broken down computers. That is a fairy story for people afraid of the dark." ~Stephen Hawking
  Reply With Quote
Old 15th August 2009, 04:42 AM   #5
iko is offline iko  Canada
diyAudio Moderator
 
iko's Avatar
 
Join Date: May 2008
Location: Toronto
You must know the unit!

Did some poking and pulled the horizontal amp & sweep board out. Powered it with a lab supply, +8V, -8V, +100V. The voltage on the pins P2334 P2344 (sweep to crt) was skewed to positive +90V, and didn't change while turning the horizontal position pot. More poking and the problem ended up being Q2334 and Q2344, both 2N5551. Replaced them with a couple of mpsa42 that I had on hand and now the voltage on those two pins can be nicely balanced to approximately 50V each. Can't test it in the scope because I left it at work but I'm pretty sure it should work.

Thanks again for your pointers guys!
Attached Images
File Type: png scope-t922r-hrzsweep.png (20.1 KB, 93 views)
  Reply With Quote
Old 3rd May 2010, 10:12 PM   #6
mjf is offline mjf  Austria
diyAudio Member
 
Join Date: Jan 2009
Location: Austria
hello.
had the same problem with my leader scope,the horizontal trace moved to the left side and could not be moved to the right...........found out that a transistor (2sa1371) and a res (8,2k blue) in one half of the output stage of the horizontal board were damaged.......
built in new parts and it works again.
greets
Attached Images
File Type: jpg oszi2.JPG (131.2 KB, 31 views)

Last edited by mjf; 3rd May 2010 at 10:17 PM.
  Reply With Quote

Reply


Hide this!Advertise here!
Thread Tools Search this Thread
Search this Thread:

Advanced Search

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are Off
Pingbacks are Off
Refbacks are Off



New To Site? Need Help?

All times are GMT. The time now is 11:10 PM.


vBulletin Optimisation provided by vB Optimise (Pro) - vBulletin Mods & Addons Copyright © 2014 DragonByte Technologies Ltd.
Copyright 1999-2014 diyAudio

Content Relevant URLs by vBSEO 3.3.2