How to calibrate Fluke 87 meter? - diyAudio
Go Back   Home > Forums > Design & Build > Equipment & Tools

Equipment & Tools From test equipment to hand tools

Please consider donating to help us continue to serve you.

Ads on/off / Custom Title / More PMs / More album space / Advanced printing & mass image saving
Reply
 
Thread Tools Search this Thread
Old 4th August 2009, 01:14 AM   #1
gto127 is offline gto127  United States
diyAudio Member
 
Join Date: May 2007
Default How to calibrate Fluke 87 meter?

I have a 20 year old Fluke 87 meter that I feel may need calibrating due to age and due to some different measorements with milivolts compared to my scope. I'm tight on money now so I can't send it to Fluke. Could anyone tell me how to cal this thing?
  Reply With Quote
Old 4th August 2009, 01:48 AM   #2
diyAudio Member
 
astouffer's Avatar
 
Join Date: Jun 2004
Location: Pittsburgh, crumbling wasteland
The easiest way is to compare it against a meter you know is a good. I have an old Fluke 75 meter and have burned out the fuse resistor twice so far. Its only a few millivolts off from the calibrated HP 34401 meters at work. Do you know anyone with access to calibrated equipment?

DMMs will start to disagree with a scope when the sinewave has distortion or stops being a sine. Frequency response plays a big part in this too.
  Reply With Quote
Old 4th August 2009, 02:04 AM   #3
gto127 is offline gto127  United States
diyAudio Member
 
Join Date: May 2007
Thanks for reply. I don't know anyone with calibarted equip. I live in a small town. Mabye it is in spec It could be the wave forms like you said. I do have a agilent power supply which is supposed to be a very high end supply. Could this be accurate enough to do a cal or would it vary like most everything else?
  Reply With Quote
Old 4th August 2009, 02:07 AM   #4
diyAudio Moderator Emeritus
 
Iain McNeill's Avatar
 
Join Date: Oct 2007
Location: Santa Cruz, California
If you don't have access to calibrated equipment your best bet will be to calibrate to a band gap reference or failing that the Vf on an LED. If you averaged a green, red & yellow LED I dare say you'd be within 5%.

Or are you looking to calibrate the 500V scale?

AC is a different matter of course, as astouffer mentions.

I think you can do everything with a good volt reference. Anyone know other good references? What where those battery cells they used to use before NIST?
  Reply With Quote
Old 4th August 2009, 02:08 AM   #5
diyAudio Moderator Emeritus
 
Iain McNeill's Avatar
 
Join Date: Oct 2007
Location: Santa Cruz, California
BTW, I'd trust the old Fluke over everything else you mentioned. My experience with Fluke is that they are very stable.

Scopes & PSUs.........hmmm

edit: it came to me: lelanche
  Reply With Quote
Old 4th August 2009, 02:51 AM   #6
gto127 is offline gto127  United States
diyAudio Member
 
Join Date: May 2007
I was looking to cal the milivolts mainly. I'm doing tracking & focus servo adjustments on old cd players. Of course I use the scope for most of it but some measurements require a DVM. How to you use the LEDS to cal the meter?Measureing voltage across?
  Reply With Quote
Old 4th August 2009, 03:03 AM   #7
diyAudio Moderator Emeritus
 
Iain McNeill's Avatar
 
Join Date: Oct 2007
Location: Santa Cruz, California
exactly.
The forward volt drop on an LED is very stable and consistent. That's why it make such a good reference for current sources. Don't want to quote voltages from memory because i'd be wrong but they're all in the 1-2V range (except blue of course)

Knowing the rms behaviour of your meter you could use this to calibrate AC volts also.
  Reply With Quote
Old 4th August 2009, 03:54 AM   #8
diyAudio Member
 
RocketScientist's Avatar
 
Join Date: Mar 2008
Location: Northwest
You can buy a reference grade voltage reference like an LM4030, LM4132, MAX6033, REF5010, etc. But even those are typically 0.01% at best and your Fluke is supposed to be more accurate than that. I have a high-end Agilent bench DMM with a 4 figure price tag that's still under factory calibration and reads to 7 digits. But my ancient Fluke 87 agrees closely enough with it for 99% of anything I need to measure. And the Fluke has been dropped more times than I can count. So, unless you have reason to believe your Fluke is wrong, I'd just trust it.
  Reply With Quote
Old 4th August 2009, 03:56 AM   #9
gto127 is offline gto127  United States
diyAudio Member
 
Join Date: May 2007
Thanks its good to know there is a component that is consistant in vout. Even voltage regulators tend to vary in v out slightly from each other.
  Reply With Quote
Old 4th August 2009, 04:05 AM   #10
gto127 is offline gto127  United States
diyAudio Member
 
Join Date: May 2007
To Rocket scientist I was thinking there might be a problem due to some measurements i was taking the other day in the 350-400mv ranfge. The volt kept disapearing around 360mv & up. It could have been just an unstable source but I was just thinking meter may have been out of cal & I was actually in a higher volt range with source causing it to be unstable..
  Reply With Quote

Reply


Hide this!Advertise here!
Thread Tools Search this Thread
Search this Thread:

Advanced Search

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are Off
Pingbacks are Off
Refbacks are Off


Similar Threads
Thread Thread Starter Forum Replies Last Post
Fluke 8012A meter question k0uc Parts 3 9th January 2009 07:12 AM
How to adjust and calibrate BB PCM64 Mikewong Digital Source 2 10th February 2007 12:14 AM
Fluke Meter For Sale ANTHONY2181 Swap Meet 8 9th March 2003 10:40 PM
New toy: Fluke 6303a LCR meter tiroth Parts 13 4th December 2002 07:54 PM